This Topic is base on other topic in PXI Forum (Link).
The main problem is the Range passing the Min/Max limitation.
I would like to simulate a channels in "Nominal" Case.
1. The range should be based on the Min/Max limitation.
2. I could select input style (Sin Wave, Triple Wave, Random in sub range, DC with error range, based on external data,...)
3. Near to Real timing simulation include cases of sync between different channel rates.
Currently,
1. The range is limited to the Min/Max Values. Excluded the cases show below.
2. Only Sin Wave. Excluded an Odd Case show below.
3. I tried case of sync between 2 channel with different rates. the simulation is not close to the real DAQ case.
Here are the cases from the PXI Forum (Link😞
1st problematic case
1. Define a Simulated Device - PXIe-4353 in MAX.
2. Open "Test Panel"
3. Use the default parameters of the test:
a. Measurement type - thermocouple
b. Max Input Limit - 100
c. Min Input Limit - 0
d. Units - Deg C
e. Thermocouple type - J
f. CJC Source - Built In
4. Click Start.
The result is Sin Wave.
The sample values range ~[-1030,-74.7], out of the Max/Min Limits.
=> BAD sample results, values lower than -273 C ?!?!?!?!
2nd Problematic Case
If I change the Max/Min Limits to other valid range like [-200,1200]
The sample values range is BAD
And, the result is some kind of Non-Sin Wave
3rd Non-temperature Case
If I change the Measurement type to Voltage.
The sample values range is based on the Max/Min Limits.
=> So it is OK
Tnx,
Raz