06-02-2009 12:01 PM
06-02-2009 03:24 PM
06-02-2009 04:24 PM
Yeah, that was my guess. On a follow-up, I've been trying to understand the +5V pin (100) on those two boards. From page 4-5 of the manual,
• (NI 6512/6513/6514/6515 Only) To supply +5 V of isolated power, use an
external power supply over +8 V on Bank 6 and Bank 7.
Does that mean that if connect up a power supply of at least 8 V with the positive terminal to P7.COMM and the negative terminal to P7.GND, then P7.+5V will be at 5 V? Same thing for P6.COMM, P6.GND, and P6.+5V?
See also page 3-26 and 4-10
06-02-2009 04:55 PM
06-03-2009 02:54 PM - edited 06-03-2009 02:55 PM
Hello Tim,
You are correct. You need to connect an external power supply over +8V to power the +5V lines on bank 6 and 7. The COM line is also referred to as VCC on the pinouts of page 3-19 (for the 6515). Please let me know if you have any additional questions. This line is typically just used for regulating a variable supply source or dividing down a power supply source to +5V.