Hey,
It looks like we need to set up triggering for the DMM. If I understand correctly, by the time the acquisition starts, the signal has almost completed its transition. Also, note that the acquisition time frame consists of several components. These are:
1) Switch Time (if switching is used)
2) ADC calibration (if enabled)
3) Auto-Zero (if enabled)
4) Settling time
In order to make the first measurement more quickly, make sure that you have not enabled ACD calibration and AutoZero. Otherwise, a little more information about how the acquisition is currently being triggered would be helpful.
Regards,
David Hall | Applications Engineering | National Instruments