5112 Vamplitude -vs- Vp-p

I have a 5401 arb in a pxi chassis dumping a +/- 1.0 V sine wave directly to a 5112 scope in the same pxi chassis through a coax.  Using the scope function for Vp-p I see 2.125V, using the function for Vamplitude I see 1.925.  What's the difference between the two functions?
 
 
0 Kudos
Message 1 of 2
(6,619 Views)
Hi Brad,
 
You can find the difference between these two measurements in the High Speed Digitizers Help under Programming >> Reference >> NI-SCOPE LabVIEW Reference Help >> Scalar Measurements.
 
The Voltage Peak to Peak measurement is:
"The maximum voltage minus the minimum voltage in volts."
 
The Voltage Amplitude is:
"The voltage high minus the voltage low in volts. "
 
Furthermore, you can find the definition for voltage high and voltage low in the same section of the help document.
 
Voltage High:
"The last-acquisition histogram method is used, where the voltage high result is the voltage of the histogram bin with the maximum number of hits above 60% of the waveform's voltage peak-to-peak value. This calculation is useful for ignoring the overshoot and preshoot on square waves. "
 
Voltage Low:
"The last-acquisition histogram method is used, where the voltage low result is the voltage of the histogram bin with the maximum number of hits below 40% of the waveform's voltage peak-to-peak value. This calculation is useful for ignoring the overshoot and preshoot on square waves. "
 
I hope this helps.  Please let me know if you have any further questions.
 
Regards,
Sean Close
Product Support Engineer: High Speed Digitizers
Message 2 of 2
(6,613 Views)