Ca't access pxi 5122 remotly

I wan't to access my pxi system remotly using remote device access.

On the server side there're daq devices pxi-6508 and pxi-5122 installed.
After launching the server Remote Device Access Server, i managed to access to pxi-6508 from the client system, but the the problem is that pxi-5122 is not listed on the "Remote Configuration" device list. So, i can't see the scope card over the network, although it functions properly locally on the server machine.

Do you have any idea?

0 Kudos
Message 1 of 3
(6,805 Views)
Hi Mursel,
 
Unfortunately, this document points out that Remote Device Access can only be used with products controlled by NI-DAQ.  Since the 5122 is controlled with NI-Scope and not NI-DAQ, RDA won't work with this device.  Among the modular instruments NI offers, the 5102 is the only device that supports RDA.
 
An alternative would be to write a program that runs on the remote device, and send data to your local device.  If you're using LabVIEW, this could be done using network-published shared variables.


Message Edited by JamesR on 01-07-2008 11:40 AM
Regards,

James R.
National Instruments
0 Kudos
Message 2 of 3
(6,785 Views)
Hi James,

Thanks for your reply and explanation.

Mursel