The 5412 has a deterministic delay from a start trigger to the beginning of analog output. The length of this delay is dependant on the interpolation factor of the digitial interpolation filter and also your sample clock frequency. With a 100Mhz sample clock and the interpolation filter disabled the delay is 43 sample clock periods + 110 ns which would be would be ~ 580 ns. The delay in the PXI trigger lines could indeed cause a bit of delay between the AWG getting the start trigger and the digitizer getting the start trigger. Depending on which slots you have each board in this propagation delay could either add or subtract from the 580 ns. Since this delay should be deterministic measuring it and specifying the value as a trigger delay to the digitizer is a ok way to solve the problem.
You can find more info on the 5412 delay in the spec sheet by going to Start -> programs -> National Instruments -> NI-FGEN -> Documentation ->Hardware Specifications -> NI 5412 Specifications