A little late I know ...
Other options:
1. NI-Scope measurment library has a built in two channel time delay measurement. This will work if you are using a NI digitizer.
2. If you are measuring the difference between two repeative waveforms then phase delay is the most accurate method. This is where you calucated the phase of each waveform and divide by 360 degrees and multiply by the frequency of the waveform. The difference of time between the two channels is your skew. Note that the accuracy of this measurement is limited by the resolution of your device. Ex. a 14bit 100 MS/s measurement device will give you far better measurements than a 2 GS/s 8bit device.