Instrument Control (GPIB, Serial, VISA, IVI)

cancel
Showing results for 
Search instead for 
Did you mean: 

GPIB crashing when running labview

Solved!
Go to solution

I'm running labview 8.6 on windows 7 with a GPIB-USB-HS which is connected to an HP 34401A multimeter. This setup is used to record pressure readings. Lately I've been having a problem where labview will suddenly freeze and stop taking pressure values. This crash happens randomly and without warning. It may be triggered by the pressure values changing very rapidly but I'm not sure. I initially suspected the computer, the drivers, or the multimeter but I have since ruled those sources out. My only conclusion now is that the GPIB is failing in some way. Is there anyway I can diagnosis the GPIB or rectify this problem? 

0 Kudos
Message 1 of 11
(4,396 Views)

I'd recommend dumping the USB-GPIB adapter and using either a PCI GPIB card, or communicating with the instrument via serial. 

 

My guess is the drivers you're using for your USB-GPIB adapter aren't playing nice with Win7.  Honestly, I haven't found a USB-GPIB adapter I like yet.... the PCI ones work well ($$), and serial is slow but works a lot more reliably.

0 Kudos
Message 2 of 11
(4,390 Views)

Hi,

 

If you installed GPIB at the time you installed LabVIEW 8.6, it would have installed NI-488.2 for Windows, Version 2.6. However, support for Windows 7 was not added until Version 2.7.2. If you have an unsupported version of GPIB on your Windows 7 system, you need to uninstall it first. Here is a snippet from the Readme.txt file that explains how to do that.

 

Unsupported NI-488.2 Versions on Windows 7

NI-488.2 for Windows, Version 2.7.2, was the first version to support Windows 7. If you installed a version of NI-488.2 lower than Version 2.7.2 on Windows 7, you must uninstall it. To uninstall on Windows 7, go to the Control Panel and select Uninstall a Program.

 

Once you have uninstalled the previous version and restarted your system, you can then upgrade to the latest driver, which is available here:

 

NI-488.2 3.0.2

 

Hope this information helps.

 

gpibtester

0 Kudos
Message 3 of 11
(4,376 Views)
If you can reliably reproduce the issue, you could run the application with NI-I/O Trace (formerly NI-Spy) running. This should let you see whether it is stuck in one of the GPIB calls when LabVIEW hangs. Jason S.
0 Kudos
Message 4 of 11
(4,359 Views)

So I ran the NI I/O trace as suggested and eventually the following error popped up.

 

> 2788. GPIB Wait (19, 4294967295, 37236)
> Process ID: 0x0000029C Thread ID: 0x00000238
> Start Time: 09:49:09.947 Call Duration 00:00:00.799
> Status: 28 (0x1C) 

 

As I suspected this address indicates that my pressure sensor is causing the problem. Does anyone know what status 28 means?

0 Kudos
Message 5 of 11
(4,321 Views)

The status is a byte that is defined by your instrument but probably bit 5 signals a srq and that should make it above 28. so try to find a definition in your instrument.

greetings from the Netherlands
0 Kudos
Message 6 of 11
(4,314 Views)

Actually in this case Status refers to the status value that is returned from the GPIB Wait function.

 

Do you have the .spy file available that could be posted, so that we could open it in NI I/O Trace for viewing?

 

-Jason S.

0 Kudos
Message 7 of 11
(4,310 Views)

So I've attached the .txt file for a similar crash. The default for my version of NI trace was a .nitrace file the posting of which this forum did not support. The program I'm running does not seem to contain GPIB write but rather is utilizing Wait for GPIB RQS. 

0 Kudos
Message 8 of 11
(4,306 Views)

I think you forgot to include the attachment. Will the forum let you post the iotrace file if you put it in a zip file first?

0 Kudos
Message 9 of 11
(4,291 Views)
Solution
Accepted by topic author lowtemp

It looks like our problem was that we had a bad connection with our gpib cable, causing labview to time out waiting to talk to the instrument it was connected to. We swapped out the bad cable and no more crashing.  Thanks!

Message 10 of 11
(4,282 Views)