I have the N6700 chassis and have installed the IVI driver for it. I have two different modules in the chassis. Both exhibit the same issue. I initialize them, which works fine. I disable the outputs, and then query the max voltage level. All this works and retunrs good data. Then I try tyo query the max current. I have tried with both the IVI function as well as the driver specific function. Both timeout. I have tried passing 0V, 20V, and the output from the max voltage back into its voltage input. It doesn't like the 0V, the other two options just timeout. I use this same routine to querya couple E363x supplies and it works fine on them. Any thoughts on why this specific VI isn't working?
Solved! Go to Solution.
07-10-2014 08:39 AM
Does the instrument have any errors after the timeout? Make sure you do a querry for all of the errors from the instrument. That will usually key you into the problem.
07-10-2014 09:08 AM
I get two errors.
GPIB: Error -101, Invalid character
GPIB: Error -420, query unterminated
Is there something wrong with the IVI driver?????
07-10-2014 09:23 AM
OK, it was the driver again. I am not going to download any more drivers off NI's website. That is the third bad driver I have gotten from them. Got the correct one off Agilents website and it works fine.