A few things I would recommend.
1. Use ungrounded thermocouples. I've seen grounded thermocouples cause ground loop problems in the past.
2. Use shielded thermocouple and ground only one end of the shield.
3. Acquire the thermocouple data at a much higher rate and average the results
08-31-2009 04:39 PM
I agree with everything ravans fan said... Also, isolated inputs is a good idea.