Agilent 3634 power supply (Timeout error)

Solved!
Go to solution

I try to send *IDN?\n query in VISA test panel( i have connected Agilent 3634 PS through RS-232 port), but it is throwing an error: 

Read Operation
Error

VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.

I am able to do Write operation but not read. Can anyone explain what the issue might be here.

tan0709_0-1710411770270.png

 

0 Kudos
Message 1 of 10
(791 Views)

try setting to read just 6 bytes and see if it reads the data. Right now you're telling the program to wait for 1024 bytes and it gives a time error because the 1024 bytes didn't come. 

0 Kudos
Message 2 of 10
(756 Views)

@tan0709 wrote:

I try to send *IDN?\n query in VISA test panel( i have connected Agilent 3634 PS through RS-232 port), but it is throwing an error: 

Read Operation
Error

VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.

I am able to do Write operation but not read. Can anyone explain what the issue might be here.

tan0709_0-1710411770270.png

 


Is your serial protocol set correctly?

mx1Capture.PNG

 

Is Flow Control needed and set correctly?

mx2Capture.PNG

 

The I/O settings are even more important

  1. Try increasing  the timeout
  2. Does your device send a Termination Character?
  3. I/O Protocol Normal or 488.2

mx4Capture.PNG 

 

Most importantly is your device IEEE 488.2 compliant?

If it is not compliant then it will not respond to IEEE Common Commands like *IDN?

========================
=== Engineer Ambiguously ===
========================
0 Kudos
Message 3 of 10
(748 Views)

Timeout error on serial devices is quite frustrating because it can mean any of the following:

 

Device is off

Cable isn't plugged in

Device is on the wrong port

Your USB to serial adapter is bad

Cabling is all correct but your cable is bad

Cabling is all correct but you are using the wrong baud rate or other RS232 setting

Everything is set up right and connected right but your device has locked up internally

Everything is connected and working properly but you sent it a command it doesn't recognize or doesn't think it needs to reply to

 

As such there's no single magic bullet solution.  I often start off by trying to use the manufacturer's software if they have some, and try connecting to every COM port.  Doing that usually means that you can isolate the problem to a much smaller set of those potential problems, since the manufacturer's software removes the chances of using the wrong port settings, wrong queries, and so on.  

0 Kudos
Message 4 of 10
(721 Views)

Changed the byte size still giving the same error.

0 Kudos
Message 5 of 10
(704 Views)

Thank you for the detailed guidance. But still none of it worked. I tried:

1. Changing Data rate settings according to the manual.

2. Tried with changing buffer size too

3. Increased timeout, with termination character and with I/O 488.2 protocol also

 

I think i have tried all the combinations now, but still if you think there could be any other reason why it is not working.

 

tan0709_0-1710483132514.png

does this has to do anything with this warning? 

0 Kudos
Message 6 of 10
(701 Views)

Hello Kyle,

 

Regarding the reasons you have stated, I will go one by one.

 

1. Device is off-- I assure you it's not OFF.

2. Cable isn't plugged in -- correctly plugged.

3. Device is on the wrong port. -- I don't think this is the case.

4. Your USB to serial adapter is bad - so, I am able to send commands like *RST\n and it's working fine.

5. Cabling is all correct, but your cable is bad -- since I am able to run some commands, I don't think its related to the cable.

6. Cabling is all correct, but you are using the wrong baud rate or other RS232 setting -- I am currently using default Baud rate setting as mentioned in Manual, although I have tried with all possible combinations.

7. Everything is set up right and connected right but your device has locked up internally.

Everything is connected and working properly but you sent it a command it doesn't recognize or doesn't think it needs to reply to. --refer the image (are you talking about the same thing?)

 

tan0709_0-1710484156874.png

it was previously in Acquire Exclusive lock and when I changed it to Release Lock it showed this error.

0 Kudos
Message 7 of 10
(696 Views)

@tan0709 wrote:

 

I think i have tried all the combinations now, but still if you think there could be any other reason why it is not working.

 


The manual should give you all the proper serial protocol settings.

 

Is the instruments set to "remote" operation mode? Some instruments have to be manually set to operate remotely.

========================
=== Engineer Ambiguously ===
========================
0 Kudos
Message 8 of 10
(674 Views)
Solution
Accepted by tan0709

I went through the manual again and noticed this connection. The pin configuration of my cable was not similar to the manual (checked the wire configuration using multimeter). So, I made a new wire with same configuration as it was mentioned in manual. IT WORKED!!!!

tan0709_1-1711456818716.png

 

0 Kudos
Message 9 of 10
(571 Views)

If it works, that implies that you either need flow control or that your wiring had TX(3) to TX(3) and RX(2) to RX(2) rather than TX(3) to RX(2) like the corrected cable (Crossover) would correct.