I'm trying to configure a VI for a student lab project reading up to a dozen thermocouples. I've been told (by a source of unknown reliability) that I can wire the TC's directly to the inputs of our USB-6251 (as opposed to using an SCXI chassis with TC module). I tried running a task this morning with the following settings:
Type K thermocouples
Constant CJC at 25C
Input range 0-150C
The TC's are exposed to temperatures in the 40-50C range, but are reading temps of 800-1000C.
Is there a way to correct this, or will we be forced to dedicate our SCXI-1000/1102 combination to this apparatus?
10-09-2016 06:49 PM
@deadstick wrote:I'm trying to configure a VI for a student lab project reading up to a dozen thermocouples. I've been told (by a source of unknown reliability) that I can wire the TC's directly to the inputs of our USB-6251 (as opposed to using an SCXI chassis with TC module). ....Is there a way to correct this?
Yes, find sources of known reliability. The correct connection depends on the USB-6251 configuration. If you have the screw-terminal termination flavor you are good to wire the thermocouple wires directly. Mass terminated devices need an external Signal conditioning device. BNC terminated devices may not support thermocouple measurements.
Now you ask why! The Screw terminal type has a cold junction reference at the terminals. The mass terminated version does not have a CJC- it expes=cts a CJC at the breakout end of the cableing. A BNC connector often involves nickle plated aluminum outer conductors and copper, or gold plating on the inner conductors (all sorts of bi-metal junctions.)
From your post I'm betting that the metalicity of the thermocouple is not maintained to the cold junction.
10-09-2016 09:39 PM
10-12-2016 10:02 AM
10-12-2016 10:20 AM
Are you configured for differential inputs? If you think you have everything set up correctly post a daigram of your connections and your VI.