10-03-2023 10:22 PM
10-03-2023 10:24 PM
Does GL820 support IDN query?
10-03-2023 10:48 PM
Hi, sorry i new the forum. it does work from NI MAX
But error start from VISA read block
10-03-2023 11:32 PM
So, you need to watch this first - https://labviewwiki.org/wiki/VIWeek_2020/Proper_way_to_communicate_over_serial
10-04-2023 06:51 AM
I am not finding a programmer's manual for your device (I didn't spend much time searching either). So I will trust that your command it correct. However, you likely need to concatenate an End Of Line character(s) to your command string before it goes into the VISA Write. Instruments tend to look for that Carriage Return and/or Line Feed to know when the command is complete.
10-05-2023 01:40 AM
10-05-2023 02:16 AM
Your write string is not correct.
Are you sure you are adding a New Line or Carriage Return to the end of the string?
No one is able to debug your picture.
10-05-2023 07:12 AM
Change the display on your command string constant to '\ Codes'. It also helps to make the display style visible (right-click->Visable Items->Display Style"). By default, the display style is 'normal'. So you are not sending a Carriage Return and Line Feed, but literally '\', 'r', '\', 'n'.
10-05-2023 07:31 AM
Agilent Technologies / Keysight Technologies 34970A Data Acquisition Unit - Instrument Driver is readily available.