Yes excatly, I only have this problem with Labview ( the PCI card infact). The signals are ok when red with a multimeter. I don't know if I was clear, but the PCI card and Labview do the correct measure if the signal is red first ( MAX, Daq Assistant, and DAQmx say the same).
That's why I think there is a problem with the multiplex when changing impedance at a "high" speed....... I haven't got this problem if I mesaure 2 voltages from 2 power supplies ( which have the same impedance I suppose).
I have tried with another computer and another PCI 6221 card, and it is exactly the same.
I also did an auto calibration. Nothing changed.
07-10-2007 02:33 PM
07-10-2007 02:56 PM
07-10-2007 03:01 PM
I had a similar problem and to solve it, I just add a delay to be sure that the buffer had time to refresh at least one. after that my problem was gone. The delay was about 10 second. I had a huge buffer and a sample time quite slow.
Benoit Séguin
07-10-2007 04:22 PM
You may want to re-post thsi to the DAQ forum.
Typically it can be fixed by setting the inter-channel delay or using a buffer (i.e. Op amp gain =1) so that the input cap. can charge.
Sorry about the lack of details, but these hardware inteface issue get sticky and complicated and as I am a volunteer, I can't spend to much time working out the details.
Still trying to help,
Ben