I have my unit and PC connected between an ethernet rj-45 cable. The Yokogawa MX100 software detects the unit so there is a connection. Should I still need a crossover cable?
After running the Getting Started.vi, it brought me to the Front Panel window of the Yokogawa. The Run button showed the 'broken run' icon, so therefore it had errors. Most of the errors has the description, "The subVI is not executable. You must fix all errors in the subVI before this VI can run."
What do you think can fix this error?
12-15-2008 12:46 PM
12-15-2008 02:06 PM
Yes, it's the Getting Started.vi that has the broken run arrow. Where do you have your YKMX100 folder located. Labview may be trying to retreive the subVIs from a specific location but I don't have it there.
12-15-2008 02:09 PM
12-16-2008 08:25 AM
Hey SJPVUL,
Have you been able to successfully open your Getting Started.vi with Matthew's provided information? If not please post back with more information about what you've tried to open the VI so we can continue to help.
Thanks,
12-18-2008 11:01 AM
Thanks for your help Matt and thanks for the follow up David. I was able to connect my Yokogawa MX100 to the Labview v8.5. Labview could not automatically find the files in the YKMX100U.llb file so I manually search for the missing files and it was able to load correctly. After I was able to run the Yokogawa VI, I saved it on my desktop and now I'm able to reconnect with ease by clicking on the desktop icon.
Thanks again for the great help support! Happy Holidays!