10-14-2013 09:10 AM
The transducer is an IEPE stype and when I turn on IEPE excitation, I can't do a DC coupling. Am I left to using some type of external amp?
10-15-2013 04:32 PM
http://zone.ni.com/reference/en-XX/help/370466V-01/mxdevconsid/iepe/
"A DC offset is generated equal to the product of the excitation current and sensor impedance when IEPE signal conditioning is enabled. To remove the unwanted offset, you should enable AC coupling. Using DC coupling with IEPE excitation enabled is appropriate only if the offset does not exceed the voltage range of the channel."
Hope this helps!
-Eric E.
10-16-2013 07:17 PM
The 9233 module only allows AC coupling and the 9234 will not allow DC coupling and IEPE excitation at the same time.
Thanks for your response.
10-17-2013 10:58 AM
I believe this may so that we maintain the signal within the specified range of the device. Without AC coupling we could see a high signal that is outside of the range of the device.
Thanks!
-Eric E.
10-17-2013 11:44 AM
From memory, PCB ICP load cells (that's IEPE by another name) have a typical bias voltage of around 10 V, which is usually stripped by a capacitor on the excitation amplifier (or the 9234 in this case). I used to have a series of triaxial load cells that were biased to 10.25 V.
PCB sell a number of relatively inexpensive external amplifiers themselves, but take care to mention to their sales representative that your interest lies in static or quasi-static measurements. I bought a 482C05 four channel amp and decided to use it for low-rate tests, and found that the decay was considerable in one set of tests thanks to the amplifier time constant. One cheap and cheerful solution was to connect a BNC T-piece at the amplifier/load cell connector and branch off the signal. This is where the bias voltage is important - I was fortunate to have a 16 bit DAQ with which I could just about measure signals using the 10 V range - as you would need to remove the DC offset somehow to read with a 5 V DAQ. I was planning on using a nice steady voltage regulator opposing the load cell until I found my DAQ was capable of handling the measurement. On the plus side, DC performance was now no longer affected by the amplifier's frequency response 🙂
So, tl;dr - get another excitation source and remove the DC offset voltage if you want to measure in DC coupling.