Im using Agilent 34970A for some voltage measurement purpose, Im getting error 521 and 103 when i select more than 15 channels.
I understand that Error 521 is an input buffer overflow error. Im using one of my own VIs to do the measurements.
I have measured 40 channels at a time and i never got an error, but when i use the VI attached im getting the error.
I did refer to some of the threads that deal with the same problem, i did not find a solution.
Thank you,
Eureka
Solved! Go to Solution.
11-29-2011 11:51 AM
Hello Eureka,
Can you perform a highlighted execution to find out exactly which VI is generating the different errors you are getting?
Mart G
Applications Engineering
National Instruments Corp.
11-29-2011 12:07 PM
Hello again Eureka,
Additionally you can check this post too:
Error 511 & 103 in Agilent 34970A
Regards,
Mart G
Aplications Engineer
National Instruments Corp.
11-29-2011 04:16 PM
Thank you for the reply...i figured it out...had to configure the instrument to XON/XOFF mode so that it keeps feeding the input channel values. I did not notice that i made changes in the configuration while i was making someother changes...
@margasan,
The post you directed me to was one of my own posts...but thank you anyway...
Thank you,
Eureka
11-29-2011 04:20 PM
Ohh sorry about that, I did not check the user in that post! So you have seen the same error in different situations!
Well I am glad to hear that the problem got solved
Regards
Mart G
11-29-2011 05:39 PM
margasan,
Not a problem,
Yes, I saw the same error in different situations. Its surprising how simple the solution is
Im pretty sure im gonna be seeing many more errors but i just hope they'll be different errors
Thank you,
Eureka