LabVIEW

cancel
Showing results for 
Search instead for 
Did you mean: 

Load Cell configuration

Solved!
Go to solution

I didn't ask for the power requirement  of your sensor 🙂

I asked for the requirements of your force measurement 😉

 

Bridges can be measured with any a wide range of supply voltages (with some exceptions.. piezoelectric bridges, accuracy, self heating, ...)  

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 11 of 18
(844 Views)

Hi all,

 

I am trying to calibrate an actuator load cell (interface) with 200 kip limit to an external NI cDAQ Module NI 9215 BNC. I just don'y know which scale factor to use for the conversion of voltage to lbs or kips since the documentation from the Load cell is in mV/V. Any Idea how to perform this calibration?

 

Thanks

0 Kudos
Message 12 of 18
(799 Views)

Hi all,

 

I am trying to get a reading from an actuator's Load Cell (Interface) through my cDAQ and Module NI 9215 BNC. The output signal from the actuator is through a BNC cable as shown in the picture attached. I am not sure how much excitation is applied in order to calibrate the readings on the range of -/+ 10 V. The values from the load cell calibration document are in mV/V, how can I transform that to a range of +/- 10 V

 

Thanks

 

0 Kudos
Message 13 of 18
(825 Views)

Hi hassan,

 

there are surely more data in the datasheet than just this single mV/V value...

 

(What kind of unit is "kips", never heard of before!?)

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 14 of 18
(793 Views)

Hi Hassan,

 

please stick with one thread!

 


@askhassan1 wrote:

I am not sure how much excitation is applied in order to calibrate the readings on the range of -/+ 10 V. The values from the load cell calibration document are in mV/V, how can I transform that to a range of +/- 10 V


As I already answered to your other message: the datasheet of your sensor should provide more information than this single "mV/V" value…

Mind to attach that datasheet?

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 15 of 18
(783 Views)

Thanks for your reply !

 

Attached the load cell original documentation.

 

Thanks!

0 Kudos
Message 16 of 18
(778 Views)

Hi hassan,

 

Why do you want to scale to +-10V at all?

 

You should know the supply/excitation voltage as the NI9215 is a simple voltage measurement device. Then you can divide your signal by supply voltage to get the mV/V value as provided in the calibration sheet!

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 17 of 18
(757 Views)

Hi, 
I saw your answer about the connection between the load cell and the NI 9215. And I have a similar problem. 
I am trying to connect a load cell (4 wire load cell) to a either a NI 9220 or NI9264. Also I have a Pandajojo JY-S65 load cell senor transmitter amplifier module. It is my first time trying to put a circuit together. I would appreciate your help

0 Kudos
Message 18 of 18
(82 Views)