12-30-2021 05:48 AM
Hallo All
I am trying to measure the Current voltage characteristics of 4 solar panels. Its just that when I connect solar panel 1 to one channel and solar panel2 to channel 2 and reverse the connections I dont read the same values. I am not able to trouble shoot this problem. Any help is much appreciated.
12-30-2021 07:48 AM
Hi Codyzded,
@Codyzded31 wrote:
I am trying to measure the Current voltage characteristics of 4 solar panels. Its just that when I connect solar panel 1 to one channel and solar panel2 to channel 2 and reverse the connections I dont read the same values. I am not able to trouble shoot this problem.
12-30-2021 12:46 PM
Hi Gerd,
I am using NI 9205. I have totally 13 channels ( Voltage: 4, Current:4, Temperature: 4 and speed:1 ). I have still not had the oppurtunity to measure with voltage and current of all 4 panels( waiting for a part delivery). I have measured with one panel current, voltage and temperature. Found the values to be ok. Current and voltage confirmed with multimeter. Temperature used a solder gun with known input temperature and showed correct results.
Please let me explain: So first I connected channel one to solar panel one and channel two to solar panel two ( differential measurement). Then I put solar panel two to channel one and solar panel 1 to channel 2. This is what I meant by reversing. I have also only isolated both these channels that means I am not connecting and reading anything else on the DAQ. Just two voltages from two solar panels. I tried adding a dummy channel but this doesnt solve my problem.
Any more help will be appreciated. Thank you.
12-30-2021 01:11 PM
Hi Codyzded,
so you are using a NI9205 module?
@Codyzded31 wrote:
So first I connected channel one to solar panel one and channel two to solar panel two ( differential measurement). Then I put solar panel two to channel one and solar panel 1 to channel 2. This is what I meant by reversing.
So we are talking about pins AI0, AI1, AI8 and AI9? And you switched all of them?
12-31-2021 07:36 AM
Yes. I do also want to tell you that the DAQ and the source are connected to the same ground. But the voltages I am getting are not from the same ground. I mean I have 3 megaohms in series ( 1 X 3 times) and i am taking the voltage reading from across one of the resistor. But solar panel is connected across 3 mega ohms. The negative of the solar panel, one end of the 3 megaohm are all connected to the common ground.
I also want to mention I have no resistors for bias currents. But I am not sure if it is a grounding Problem..
Please want to be clear: A0 and A8 : solar panel 1 : A1 & A9 Solar panel 2.
2nd try: A1 & A9: solar Panel 1; A0 & A8 : solar panel 2.
sorry for the late reply only have 2 posts per day as new user..
01-04-2022 03:31 AM
Hallo All
Just an update on my problem. I am measuring only 4 voltages of 4 solar panels. I have connected the first 4 channels for solar panel 1-4. I got a set of values and then I changed only solar panel one and solar panel 4 to each others channels respectively. I dont get the exact same set of values. Any help will be appreciated.
01-04-2022 12:00 PM - edited 01-04-2022 12:01 PM
Your description of your electrical connections is confusing, please post a detailed schematic diagram of all your connections in both the configuration that worked and the configuration that failed.
01-04-2022 12:48 PM
@Codyzded31 wrote:
I do also want to tell you that the DAQ and the source are connected to the same ground. But the voltages I am getting are not from the same ground. I mean I have 3 megaohms in series ( 1 X 3 times) and i am taking the voltage reading from across one of the resistor. But solar panel is connected across 3 mega ohms. The negative of the solar panel, one end of the 3 megaohm are all connected to the common ground.
I agree with RTSLVU, your verbal description is not making sense. My first guess is that the statement I highlighted is probably the reason you are not getting the same values. But again, please provide a detailed electrical schematic.
01-05-2022 03:08 PM
Hallo All
Please find the figure below attached. The first and second case the values must match but they dont.
01-05-2022 04:14 PM - edited 01-05-2022 04:15 PM
That is not a detailed schematic of your electrical connections.
I need to see how your entire system is wired, including the three one meg ohm resistors in both configurations!