Ocean Optics USB 2000+ not recognozed by Ni-MAX

Solved!
Go to solution

I have an Ocean Optics USB 2000+ that I'm trying to use with LabVIEW 2018 on a Windows 11 PC.  I need to get the VISA resource associated with the device, but the device isn't recognized by Ni-MAX. The spectrometer works just fine when using the Ocean View software.  I've tried downloading and installing the device drivers from NI.  I get and "not enough memory" error when trying to install the .inf file.  Not sure what to do next.  The NI device driver lists the very model I have.  I don't think it's a wrong driver problem, but I suppose it could be.  The device is recognized by Ocean Optics Ocean View software and works just fine, but I cannot get it recognized in LabVIEW.  The install directions that were provided are attached.

0 Kudos
Message 1 of 2
(319 Views)
Solution
Accepted by topic author Jim12345678

I followed the attached directions

 

https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019La2SAE&l=en-US

 

and replaced the divers that were installed by Ocean View and reported as the most appropriate with the ones provided by when I downloaded the drivers from NI.