09-11-2019 05:00 PM
Hey guys,
I'm running astronics PXIe 1209 as function generator and PXI 5172 as scope on NI PXI 1092 chassis.
I fond that the time clock of fungen and scope is different; they are using their own VCXO.
do you know how to synchronize them? (with backplane 10MHZ clock?)
I appreciate your time very much!
Thanks!
10-11-2019 08:59 PM
Apply a 10 MHz reference to one of the front panel inputs of the PXIe-1209. Define that input as a reference clock input using the instrument driver. That 10 MHz will now be what the PXIe-1209 channel is referenced to. The pulse generator clock should not drift from the timing of another instrument such as a scope that is referenced to the same clock.