@Kevin_Price wrote:

This is actually true for option 3 as well.  No physical wiring terminal is needed when you use the internally routed "_ao0_vs_aognd" channel for the AI task.  You still use up the AI subsystem though.

 

 

-Kevin P


But with Option 3 the dummy AI task still take up a physical DAQ AI input pin, correct? In other words that, no physical wiring or "pin" but the AI channel would no longer be available for other purposes/tasks elsewhere in my application?

^TeraTech.

CAUTION! New LabVIEW adopters -- it's too late for me, but you *can* save yourself. The new subscription policy for LabVIEW puts NI's hand in your wallet for the rest of your working life. Are you sure you're *that* dedicated to LabVIEW? (Check out this lengthy post of mostly complaints)
0 Kudos
Message 11 of 12
(224 Views)

Correct.  In fact, since you're using a multiplexing device, any use of AI will reserve the whole AI subsystem, not just the individual channel.  There's only 1 converter and access can only be shared via multiplexing when channels are part of the same task.

 

So if you want to preserve the AI subsytem for other things, you should use option 1 and generate a trigger signal with either a DO line or a different Ctr output.

 

 

-Kevin P