05-31-2016 09:00 AM
Hello,
i am using an Agilent 33210A function generator and i try to generate a simple sinus sweep with constant amplitude.
For some reason the amplitude is increasing for the whole sweep-time. I basically used the Sweep VI from the Agilent 33xxx example VIs.
I only added the VOLT command to control the amplitude as described in the User's guide. Hope you can tell me where my mistake is. 🙂
Solved! Go to Solution.
05-31-2016 10:56 AM
Does this phenomenon occur if you sweep the frequency manually?
05-31-2016 03:06 PM
@Klüs wrote:Hello,
i am using an Agilent 33210A function generator and i try to generate a simple sinus sweep with constant amplitude.
For some reason the amplitude is increasing for the whole sweep-time. I basically used the Sweep VI from the Agilent 33xxx example VIs.
I only added the VOLT command to control the amplitude as described in the User's guide. Hope you can tell me where my mistake is. 🙂
Have you tried setting the amplitude before configuring the sweep, not in the middle of it?
Ben64
06-01-2016 01:54 AM
Thank you, that was the right hint. It was not a software problem but a hardware thing. I didn't consider that the measuring chain has such a frequency depending behaviour.
Obviously a beginner's mistake. Next time i will try to exclude these types of errors before i ask for help.