I have current Labwindow/CVI with Teststand. When I run Ag34980A_INIT command to initalized the AG34980A, I'm getting IVI error code -2147483648. The AG34980A communicates without a problem if I use NIMAX software. Can you help why I'm getting this error?
Thanks,
11-02-2016 07:21 PM
Hi,
Is this issue happening when you call the code module from TestStand or when you try to run the code on CVI? Are you adding the correct DLL and libraries on CVI?
Regards.
11-03-2016 07:34 AM
thank you. I did not have .DLL in proper location.