Hi
I am using LabWindows to remote-contorl the CMW 500.
I am using WCDMA Drive from the R&S to run CMW 500.
When I am turn Multi Evaluation mode to see the result of test.
it keeps displaying "Resource Conflict"
I am not really sure about reasons that I am getting "Resource Conflict"
I do not believe that I put wrong inputs...( or i am calling the extra function by LabWindows to control the CMW 500?)
is there any possiable reason that I made mistake in LabWindows ?
thanks
Solved! Go to Solution.
08-02-2013 08:58 AM
Hi pc486,
Is it possible that the hardware resource is already reserved and we are trying to access it in multiple places? Would you be able to post a screenshot of the error code? LabWindows/CVI does not have any extra functions that are called, as it runs standard C code. Also, I have found a driver in our Instrument Driver Network that you may find helpful. This was not written nor supported by National Instruments, but it may be helpful.
CMW500: http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=15501
Regards,
Jason D
Applications Engineer
National Instruments
08-02-2013 11:40 AM
Thanks for email me
I just fingered out the reasons.
I accidently set the wrong option to remote CMW500 by LabWindows.
I am guessing that calling extra function becomes resource conflict issues.
thanks for helping 🙂