-200284 error message when calling DAQmxReadAnalogF64() using a USB-6210

Solved!
Go to solution

I am trying code a CVI application to communicate with a USB-6210. I want to start the acquisition to acquire 30 samples and then collect them. My problem is when I call the function DAQmxReadAnalogF64() I get this -200284 error message. In a nut shell the error message means the data has not been acquired yet. It says I should make sure my start trigger is configured properly. I was trying to use the internal clock to start the acquistion. I figure I did not set up something properly.

 

In the function DAQmxCfgSampClkTiming() I set the sampleMode to DAQmx_Val_FiniteSamps which I thought would make it collect the 30 samples (number_of_scans=30) when I called DAQmxStartTask(). That does not work. When I set sampleMode to DAQmx_Val_ContSamps I can collect the data. Below is a stripped down essential of my code:

 

 returnAcq = DAQmxCreateTask ("myAcqTask", &acq_taskHandle);
 returnAcq = DAQmxCreateAIVoltageChan(acq_taskHandle,physicalChannel, "acq_channel",DAQmx_Val_Diff, minVal, maxVal, DAQmx_Val_Volts,"");

 returnAcq = DAQmxCfgSampClkTiming( acq_taskHandle, "OnboardClock", sweepsInterval,DAQmx_Val_Rising, DAQmx_Val_FiniteSamps, number_of_scans);
 returnAcq = DAQmxStartTask (acq_taskHandle); 
 returnAcq = DAQmxReadAnalogF64(acq_taskHandle, -1, 5, DAQmx_Val_GroupByChannel, Data_readings, number_of_scans, &read,0);
 // Clear the task
 DAQmxClearTask (acq_taskHandle);

I guess my question is: Am I set up so the internal clock will start the acquistion when I call DAQmxStartTask()? I thought when I set the second parameter in DAQmxCfgSampClkTiming() to  "OnboardClock" this would make it use the internal clock.

 

Thanks in advance for your help,

Don Pearce

 

0 Kudos
Message 1 of 3
(3,238 Views)
Solution
Accepted by topic author DPearce

Don,

 

You are correct, your task is configured to be using an internal sample clock with a software start trigger.  The error description points out several common reasons why people might see this error.  What is the value of 'sweepsInterval'?  I believe that as configured DAQmx should be clocking data in at the rate (in Hz) specified by sweepsInterval.  By passing -1 into the numSampsPerChan input, I believe that DAQmx will attempt to read all samples for a finite acquisition (30 in your case).  If you are running less than 6 Hz, the read could would timeout before this data was available.

 

When configured as continuous, what values is passed back in the sampsPerChanRead parameter?

 

Dan

0 Kudos
Message 2 of 3
(3,234 Views)

I looked at it again and realized that sweepsInterval was wrong; it was in sample time instead of sample rate. I wanted to take a sample every 1.67 ms and I had sweepsInterval= 0.00167. So when I changed that input to 1.0/sweepsInterval everything was ok. Looks like it was the ole loose nut behind the keyboard problem. Thanks for you help.

0 Kudos
Message 3 of 3
(3,225 Views)