Custom Scaling Error 200077

Solved!
Go to solution

Hello,

I have a cDAQ-9188 chassis and seven modules. My VI consists of a DAQ-Assistant to acquire signals. With this I created a channel "Luftdrucksensor". For the Calibration I used a Custom Scaling in which I created a characterisitc curve. My sensor has an output range of 4-20 mA and 0,9-1,1 bar. But when I want to start the measurement an error occurs.

Can you help me?

 

Thank You

Regards

 

0 Kudos
Message 1 of 5
(4,478 Views)
Solution
Accepted by topic author DanielDerer
The min and max must be in terms of the scaled value, not the 4-20ma unscaled measurement.
0 Kudos
Message 2 of 5
(4,470 Views)

Hello,

thank you very much, it was successful.

Just for the understanding, why must there the scaled values? In my task the other pressure channels can handle with a range from 4-20mA, whats the difference?

 

Thank You

Regards

0 Kudos
Message 3 of 5
(4,447 Views)
I suppose that technically, the limits could be installed but for ease of use, readability, consistency, I makes more sense that the limits are referenced to the scale.
0 Kudos
Message 4 of 5
(4,432 Views)
Hello Recently I have a same question,I have a cDQA-9178 chassis and NI 9213 modules,and I want to acquire Voltage signals,when I click save and run the dialog show:Error -200077,then I search a example: Voltage and Thermocouple in a single Task.vi in Labview software and try again. it also gave me a one ERROR dialog.can you help me ? Thank you very much!