Hi..
*Iam Using the Fluke Ecg PS410 Simulator as a Heart Simulator..
*Iam Using the SCB-68 Board with PCI/PXI 6225 Card for data Acquisition.
Motive:
1) I wish to Acquire the RA Signal,LA Signal and RL Signal and implement the following in LabView
Lead II = LA - RA
Actual Lead II = Lead II - RL
2) There is a High Level Connector on the Simulator at which the Lead II Signal is available so i connected an Oscilloscope with Negative pin on the RL pin and the Signal was Correct So all i have to do is to Acquire the three Signals now.
Problem:
As there is no reference for the RA , LA and RL Signals How can i Acuquire them..??
I want to know the Switch Settings and any Signal Conditioning that i might need since the signals are in the range of MiliVolts,Configurations that should be done in the NI Max Explorer and the Acquiring Option i.e RSE,NSRE,Diff etc....
Hamza Khan
07-30-2013 09:24 AM
Hi Hamza Khan,
I'm attaching a link below to the M Series User Manual. Chapter 4 on Analog Input has good descriptions of how to wire up various signal sources for data acquisition. I'm also including a link to a white paper on general field wiring considerations for analog signals.
M Series User Manual
http://www.ni.com/pdf/manuals/371022k.pdf
Field Wiring and Noise Considerations for Analog Signals