How long will the linear encoder count before rolling over?

I am measuring the angular speed of a rotating shaft with a linear encoder setting. I had expected this counter to reset at 16bits, but does not appear to have done so. How high can the counter go before it resets?

 

I will be out of the office next week. Please copy qiaoqiao.wang@voith.com on any replies.

 

Thank you.

Eric

0 Kudos
Message 1 of 5
(3,194 Views)

Eric,

 

Thank you for using the forums.  You are correct the counter should reset at 16 bits which would bed 2^16 -1 which is 65535.  What are you seeing the counter reset at?

 

Another good test for this would be a test panel for the device and do edge counting there.  Using the 20 or 80 Mhz clock you can count this very quickly to see where the rollover occurs.

 

Finally, what device are you using?

 

Regards,

 

Brian P.

Applications Engineer
National Instruments
0 Kudos
Message 2 of 5
(3,170 Views)

To the original poster--

 

What hardware are you using?  Almost all of NI's DAQ products have higher than 16-bit counters.  For example:

 

32-bit Counters:

M Series (62xx)

X Series (63xx)

TIO (660x, 6624)

CompactDAQ

 

24-bit Counters:
E series (60xxE)

 

 

The only products that have 16-bit counters are very old pre-E Series DAQ cards that are have been EOL'ed by NI.  Like E Series, these devices wouldn't support direct quadrature encoder measurements anyway, so if I had to guess you're probably using a device with a 32-bit counter.

 

 

Best Regards,

John Passiak
Message 3 of 5
(3,165 Views)

Encoders have a number of pulses per revolution (linear: per mm) . The quadratur encoder (the name give a hint 😉 ) will count 4 times per pulse (since is uses every edge of the two 90° shifted outputs).

So the time to a counter rollover is T_minutes=2^(counterbits-1)/RPM*4*<pulse/rev> 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 4 of 5
(3,135 Views)

Henrik is correct, but if you'd like you can set the mode to something other than the (default) X4 mode:

 

2011-08-16_120305.png

 

The counter will still roll over whenever you hit a count of 2counterbits.  For newer NI DAQ devices, this would be 232.  The rate at which you count depends on speed and ppm of your encoder and what decoding type you are using.

 

 

Best Regards,

John Passiak
0 Kudos