12-16-2013 02:53 PM
The initial readings on select channels of my NI9222 and NI9220 DAQ modules are significantly offset from the expected 0V value that I observe (within a reasonable tolerance) on the remaining channels. My DAQ set-up is composed of 3 modules (NI 9403, NI9222, NI 9220) that are connected via a NI 9188 chassis. I am testing my set-up with all the modules connected and without hooking up any of the respective AI+/- pins. I attached a screenshot of a test that I ran on the NI 9222. Below is a list of all the initial readings for my DAQ modules. I listed the channels with the large offset voltage that are causing me concern in bold.
NI 9222:
ai0- 10.6 V
ai1- 10.6 V
ai2- 1.97 V
ai3- 0.29 V
NI 9220:
ai0: -0.002 V
ai1: -0.003 V
ai2: -0.0005 V
ai3: -0.002 V
ai4: -0.009 V
ai5: 0.0004 V
ai6: -0.0005 V
ai7: -0.001 V
ai8: -0.003 V
ai9: -0.001 V
ai10: -0.004 V
ai11: -0.004 V
ai12: -0.009 V
ai13: -10.47 V
ai14: -0.002 V
ai15: -0.010 V
Any information on why this might be happening or how I can resolve this issue would be much appreciated.
Thanks,
Max
12-17-2013 05:51 PM
Hi wmm26,
Have you tried writing a value to these values? Use a multimeter to distinguish the value going in versus the value you are observing in MAX.
Also have you tried using a mulitmeter to measure the input values?
Also have you tried resetting these devices as well?
Have you tried swapping the slots which each module is in?
12-17-2013 06:27 PM
Hey Paolo,
I was able to resolve this issue. The step that helped me was using an oscilloscope to compare measurements that I observed using the MAX. I assumed that the ground from the power supply would be used as a default internal ground. Connecting the AIO- terminal to the grounded end of my oscillascope was a sufficient reference to re-calibrate my channel readings at 0 volts. I found it odd and slightly alarming that only certain channels exhibited this feature and at different levels. Why is this the case?
Max
12-18-2013 02:20 PM
Hi Max,
It seems that the 9222's channels are floating relative to earth ground as they channels are isolated from each other.
I found it in the specification sheet found here:
http://www.ni.com/pdf/manuals/374223b.pdf
We recomend you ground these channels with the ground used for the devices in the "-" terminal as the measurements that are expected is differential.