08-16-2016 01:12 PM
I was wondering how I can find the maximum voltage difference allowable between the AI_SENSE and AI_SENSE_2 pins on the PXIe-6363?
I need to measure a group of 16 ground referenced signals, several signals referenced to approximately -8V, and a signal that is ground referenced. The -8V reference is the most negative voltage, so this shouldn't exceed the +/-10V max voltage, so I should be safe in that regard. I was planning on doing the following:
Is this setup OK?
Thanks in advance,
Seth
08-17-2016 12:49 PM
Accodring to the device specifications sheet:
Maximum working voltage for analog inputs (and sense) (signal + common mode) ±11 V of AI GND
Overvoltage protection for all analog input and sense channels
Device on ±25 V for up to two AI pins
Device off ±15 V for up to two AI pins
08-19-2016 11:14 AM
Hey all, I would also like to share a new project we've been working on to help explain NI DAQ specifications. Hopefully this helps. We'll be glad to take feedback for that page and if it helped answer this question on the link found at the top and bottom of that page.
Specifications Explained: NI Multifunction I/O (MIO) DAQ - http://www.ni.com/product-documentation/53090/en/
08-19-2016 04:04 PM
Sebas: So the specification I'm reading doesn't have the "(and sense)" part. Specifically, the document link is http://www.ni.com/pdf/manuals/374651c.pdf. I'm assuming that's the case, but I would prefer to have that verified.
Paul, thanks for the link, but unfortunately, it doesn't address the sense lines.