02-15-2011 10:29 AM
Hi,
I'm using the NI 6250 for some Analog input analysis. In addition I want to measure the current I supply to Unit under test. The current is approximately 300ma.
I thought to use a 2 ohm shunt resistor, thus to lower the supply voltage by 0.6V and not having too low voltage for measurements. I plan to adjust the supply votage to 9.6 V so that after the shunt resistor, I'll get 9V as I'm suppose to provide.
1. Any remarks or better suggestions for that?
2. I will calculate the current by measuring the voltage over the resistor. The question is shall i use 2 single inputs or a differential input?
3. Looking in Table 4-3, page 4-12 of the DAQ M Series - User Manual, I see a recommended application of a differential input voltage. There, they put 2 pull down resistor to ground. Is this what I need to do as well? Can't I just apply the 2 lines from the resistor, directly to the diff. input?
4. Am I suppose to connect the AGND to my supply GND?
Thanks
Rafi
Solved! Go to Solution.
02-16-2011 05:41 AM
Well, depending on your load and your source I would try a current measurement with a 0.2 Ohm shunt (4 wire) in the ground line with differential measurement in the +/- 100mV range.( gives you a 500mA current range)
With 9V you might use the shunt in the high side and use a differential channel.Take a look at the specs.
If your source has sensing capabilities you might even compensate the voltage drop on the shunt.
02-16-2011 06:58 AM
Hi Henrik
Thank you for your response.
Actually, I had difficulties to obtain a 9V power supply (15W). Therefore, I decided to take a 12V ps, trimmed down (by the supplier) to 9.6V. Now, all I need is to use a 2 ohm resistor with 300mA current and I get a nice 9V for the Unit under test. My PS does not have sense lines.
Another consideration in selecting the shunt resistor was not to use a too low voltages. With 2 ohms I get 0.6 Volts. My UUT can tolerate the little voltage drop since, eventually, it regulate it down to 3.3V.
I was looking for advice whether to use a differential measurement or 2 single ended measurements...and whether I need pull down resistor if I use a differential measurement.
Thanks
RK
02-17-2011 02:34 AM
In your case the diffential measurement is the way to go. You don't need any more bias current (pulldown) resistors if your PS is connected to (A)GND somewhere anyway.