11-22-2017 07:55 AM
I want to use the NI 9218 to connect to a load cell with 2mV/V output. The 9128 data sheet refers to measurement range minimum 21.7mV/V under the full bridge characteristics heading. Does this mean I need a pre-amp to increase the signal level?
Thanks
Solved! Go to Solution.
11-23-2017 02:18 AM
The 9218 input range is ~22mV/V (+-65mV input range @~3V bridge supply)
Since the resolution is quite high, you can use it for your bridge sensor.
However you should check if the bridge is a piezo resistive one (these can be sensitive to changed bridge supplies) and a valid calibration for an exitation of 3V . But you mentioned a sensitivity in mV/V ... good indicator for a bridge supply independency 😉