NI USB 6008 Voltage Offset using RSE and Diff. Measurement
Hello everyone,
I'm currently trying out the NI USB 6008 and I'm getting problems when reading an analog voltage using RSE or Differential.
So basically all I want to measure is a PWM signal (0 to 12V), which is divided by a voltage divider (by two). But instead of measuring 0V and 6V
I am measuring a constant 0.8V and about 3V.
On the digital side of the DAQ I'm giving out the pulses for the SSR... and this works just fine.
I'm connecting it this way: http://digital.ni.com/public.nsf/allkb/95CC0CB11D7DF3D18625712E000C4ABD?OpenDocument
Would apreciate any help
Best Regards
EDIT: Attached are acquired graphs
Message 1 of 8
(6,490 Views)
Accepted by SpeerResi
What is the impedance of your voltage divider?  The input impedance of the USB-6008 is not very high. If the impedance of the divider is large, it could cause the effect you are seeing.
Lynn
Message 2 of 8
(6,445 Views)
Hey Lynn,
both are about 23 MOhm.
I'm going to change them and reply afterwards.
Thanks for your reply.
Regards
Message 3 of 8
(6,440 Views)
The input impedance of the USB-6008 is 144 kohm, so your high resistance divider is definitely a problem.
Lynn
Message 4 of 8
(6,432 Views)
Hey Lynn,
sry I was wrong, those resistors only had like 500 kOhm.
But still, I changed them and now its working.
Thanks a lot.
Message 5 of 8
(6,420 Views)
Even at 500 kohm the low input impedance of the USB-6008 would have created a significant error.
Lynn
Message 6 of 8
(6,413 Views)
I'm now using 10k. Works great!
Message 7 of 8
(6,411 Views)
Another option is to buffer the signal before going into the 6008 - use an op-amp voltage follower and feed that output into the DAQ.
Input-Z of op-amps is huge, so it doesn't load the signal.
Message 8 of 8
(6,401 Views)
