I'm currently trying out the NI USB 6008 and I'm getting problems when reading an analog voltage using RSE or Differential.
So basically all I want to measure is a PWM signal (0 to 12V), which is divided by a voltage divider (by two). But instead of measuring 0V and 6V
I am measuring a constant 0.8V and about 3V.
On the digital side of the DAQ I'm giving out the pulses for the SSR... and this works just fine.
I'm connecting it this way: http://digital.ni.com/public.nsf/allkb/95CC0CB11D7DF3D18625712E000C4ABD?OpenDocument
Would apreciate any help
Best Regards
EDIT: Attached are acquired graphs
Solved! Go to Solution.
04-16-2013 03:23 PM
What is the impedance of your voltage divider? The input impedance of the USB-6008 is not very high. If the impedance of the divider is large, it could cause the effect you are seeing.
Lynn
04-16-2013 06:24 PM - edited 04-16-2013 06:24 PM
Hey Lynn,
both are about 23 MOhm.
I'm going to change them and reply afterwards.
Thanks for your reply.
Regards
04-16-2013 10:03 PM
The input impedance of the USB-6008 is 144 kohm, so your high resistance divider is definitely a problem.
Lynn
04-17-2013 08:28 AM
Hey Lynn,
sry I was wrong, those resistors only had like 500 kOhm.
But still, I changed them and now its working.
Thanks a lot.
04-17-2013 10:05 AM
Even at 500 kohm the low input impedance of the USB-6008 would have created a significant error.
Lynn
04-17-2013 10:13 AM
I'm now using 10k. Works great!