To simplify operation of the AD595 chip use only a single operating voltage of +15VDC for Vcc. Using +15VDC will give you a larger operating range than +5VDC for positive temperatures from ambient Deg. C to +300 Deg. C. There is no need to use dual voltages unless you aer temperature cycling from a sub zero temperature to and extremely high temperature. If you are using shielded thermocouple cable do not ground the shielding because it load the AI Channel with excess noise of up to two volts on the input pin to the AD595. You will find it it better to float the ground on the shielding of the thermocouple cable to eliminate the excess noise. Also when soldering the thermocouple wires to the inputs of the AD595 will require a specific type of solder other than is commonly used for soldering. (Reference any of the three Analog Devices application notes for the solder contents.) Per NI's M-Series User's Manual states that a biasing resistor should be connected from the negative terminal to AI GND for a differential connection such as used for thermocouples. In my experience a resistor of 2.7k Ohms will work the best to eliminate noise. Also I have tried put different coupling capacitor between AI plus and negative terminals for a channel with no results in getting rid of the noise. The only thing that the different capacitance values caused was loading down the AI channel.