10-25-2023 10:06 AM
Dear all,
I would like to meassure some Current inputs 4-20 mA and some PT1000 Sensors. Is this possible with the PCIe 6323 card? There is no information about that in the handbook.
Thank you in advance!
Best regards,
Michael
Solved! Go to Solution.
10-25-2023 10:30 AM
RTD requires an excitation current to work properly. You should get a proper RTD measurement device. Measuring Temperature with Thermocouples, RTDs, and Thermistors
10-25-2023 06:10 PM
@MichaelG2 wrote:
Dear all,
I would like to meassure some Current inputs 4-20 mA and some PT1000 Sensors. Is this possible with the PCIe 6323 card? There is no information about that in the handbook.
Thank you in advance!
Best regards,
Michael
6323 is Analog Voltage input card and not best suited for the 4-20mA measurement or PT1000. You could measure it with additional external circuit but it could be extra work, instead but the right hardware for the job.
10-30-2023 07:17 AM
Is there a PCIe card that can handle a current measurement signal?
Best regards,
Michael
10-30-2023 07:48 AM
@MichaelG2 wrote:
Is there a PCIe card that can handle a current measurement signal?
Best regards,
Michael
Unfortunately, no PCIe equivalent. Here are the available models - https://www.ni.com/en-us/shop/category/current.html?page=1&foc=0
Instead, you can also use an external current-to-voltage converter.