Multifunction DAQ
PCIe-7852R and NI USB-7856R OEM Calibration with NI Calibration Executive
Can I calibrate PCIe-7852R and NI USB-7856R OEM with NI Calibration Executive?
How?
Our National Instruments Calibration Executive Software version is 3.5 and there is not procedures to products PCIe-7852R and NI USB-7856R OEM.
(4,230 Views)
Accepted by topic author antti_lappalainen
There are several ways calibrate your instruments.
Firstly, the help file for calibrating R series devices in Calibration Executive can be found by navigating to the Calibration Executive Help > Device Calibration Procedure > Data Acquisition (DAQ) Devices > NI Dynamic Signal Acquisition > NI R Series Calibration Procedure
You can also follow the instructions found in the following document.
http://www.ni.com/pdf/manuals/372004d.pdf
Lastly, you can also calibrate your software for R series devices by using the calibration utility installed with the drivers. Search for "calibrate 78xxr device" and run the executable to calibrate your devices.
Thank you.
Kind regards,
Nigel
(4,147 Views)
Hello.
Thank you for information.
Now I can calibrate R-series equipments.
Antti
(4,061 Views)
