Hello,
We have a PXI-1006 chassis with 9 DSAs (PXI-4472) in slots 2 to 10, and a MIO (PXI-6713) in slot 11. We have upgraded from NI-DAQmx 9.23 to version 9.30f2. Suddenly the 4472 in slot 2 was outside the chassis icon and the rest 8 in the chassis tree in the MAX 4.8. All 10 devices dimmed with red "x". We reset the tree and they dissaperaed however, they do not install anymore even as new devices. Not even system restart triggers their PnP discovery!
We also had the Traditional DAQ (NI-DAQ 7.44) installed, but after the above detailed event this category is not there anymore!
What sould we do - reinstall the previous NI-DAQmx, or ...? We have tried with DAQmx repiar, but at no avail...
Look forward to a solution,
Roman
04-20-2011 02:44 PM
Hello,
I would start by following the procedure described in this document. This procedure repairs Measurement & Automation Explorer in the event that the database becomes corrupted. Uninstalling DAQmx or re-installing a previous version will not correct a database corruption. Let us know if this resolves the issue.
Thank you,
Justin