Problem to identify TEDS sensors using DAQ hardware

With class 1 TEDS devices like the NI 9233, you connect the EEPROM "backwards"  Gnd -> AI+, D+ -> AI-.  With class 2 TEDS devices such as the NI 9219 and NI 9237, it is Gnd -> T- and D+ -> T+.  This is why you see different results with the serial port based on the connection.  This is in section 8 of the 1451.4 specification.  The data errors are because the data does not fit.

 

There are a couple templates that are smaller than 256 bits, but not many and you need to have disabled options.  Take a look at some of the TEDS templates.   Templates 25 and 27, for example, can be made smaller if you select case 0 for certain options (not specifying functionality).

0 Kudos