Multifunction DAQ
Thermocouple offset error w/ CJC
Hi everyone,
I am having an issue with my thermocouple reading in my PCI-6221 DAQ. I am setting up my SCB-68 accessory to use the onboard CJC compensation, and when I check my readings, it seems to be 5 or 6 degrees higher than what it suppose to be (checked with a multimeter). Here is my system info:
PCI-6221 M series DAQ
SCB-68 accessory with CJC enabled (in differential mode)
K-type thermocouple connected to AI4 and AI12 for differential reading w/ negative side shorted to AI GND
DAQmx 9.9
Labview 2010
I've checked my DAQmx parameters and things look all good. Any clue what this might be caused by? Is this a calibration issue?
Message 1 of 13
(6,463 Views)
What do. you read, when you replace the TC with a short wire?
Should be room (CJC) temperatur...
Greetings from Germany
Henrik
LV since v3.1
Message 2 of 13
(6,451 Views)
It looks like it is measuring the same temperature as my thermocouple (which was at room temperature). Warming up the onboard sensor with my hand seems to have a response. Would this be an issue with calibration then? Shoud I put in a software offset?
Message 3 of 13
(6,434 Views)
With some more tinkling and reading ("How Can I Read the Value of the Cold Junction Compensation of the SCB-68 (SCB-100)?"), I found that my cjc thermometer seems to be working okay (reading 21C, which is what I expected), but shorting a port together and to AI GND didn't yield similar results and nor did shorting it to AI SENSE.
Message 4 of 13
(6,428 Views)
Hi dragel24,
Are you checking these readings in NI-MAX or in LabVIEW?
Also, did you set your thermocouple to be K-Type in NI-MAX?
Kelly B.
Message 5 of 13
(6,402 Views)
Yes, I was using NI-MAX, and I set it to K-type in there. Also set it to use built in CJC on the SCB-68 and configured the switches.
Message 6 of 13
(6,397 Views)
dragel2014,
May I ask why you are shorting the negative terminal to AI GND? This would not be a differential measurement anymore if you do this.
Also, here are some more information in this document: http://digital.ni.com/public.nsf/allkb/96B996BAD82A84E9862570690066B03F
Message 7 of 13
(6,361 Views)
Anderson,
hmm... I have grounded the negative side to ground because it was to "prevent the differential voltage readings to drift out of the adc range". I have previously removed it, but didn't see any changes. On top of this offset, i also find my signal rather noisy. I know my thermocouple wires are a little long, but i have not see the same with another Agilent DAQ that I have.
Message 8 of 13
(6,346 Views)
What input range do you have your DAQ set on? The smallest voltage range for this device is ideal, +-0.2V, because the K-type thermocouple outputs 41 µV per °C voltage change. Any other voltage range and this 41 µV per °C would be undetectable.
Message 9 of 13
(6,316 Views)
I have it configured to use K-type thermocouple inputs. Would changing this to voltage input and doing the calculations myself help? You would think that by setting it to read K-type it would automatically configure to use the right voltage range. I have set the expected input temperature range from 0 to 100C.
Message 10 of 13
(6,313 Views)