Usb-6343 incorrect AI signal

Usb-6343 measuring 6v at terminals but daq reads .05

0 Kudos
Message 1 of 4
(857 Views)

How did you connect the signal? AI+ and AI- along with resistors to AGND? did you configure the DAQmx task to use Diff or SE based on how you connected the signal?

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
Message 2 of 4
(851 Views)

We are using a GE Druck Pressure transducer that has direct voltage out on two wires.  so we have Analog input (+) from the sensor wired to terminal 1, analog input (-) wired to terminal 2, and terminal 3 is wiring to ground. I am not aware of the need to wire in a resistor when i have voltage input, so maybe thats my problem, but all my other usage of this DAQ has had consistent DAQ reading with multimeter sampling at the DAQ terminals.

0 Kudos
Message 3 of 4
(814 Views)

See Table 1 of Field Wiring and Noise Considerations for Analog Signals

-------------------------------------------------------
Applications Engineer | TME Systems