07-22-2009 03:18 PM
07-22-2009 05:23 PM
It's simultaneous.
Valid Sampling Rate = (Master Timebase) / (256*n), where n is an integer from 1 to 31
The NI 9234 has a 13.1072 MHz Timebase, giving a maximum sampling frequency ( n = 1) of 51.2 kS/s.
The 9234 can
either use its native sampling rates or it can synchronize to one of
the other modules. So if you synchronize a 9237 and a 9234 together,
you can either set the 9234 as the master and get 51.2 kS/s for both
modules (or 25.6 kS/s or other divisors), OR you can set the 9237 as
the master and get 50.0 kS/s for both modules (or 25 kS/s, etc.).