Real-Time Measurement and Control

cancel
Showing results for 
Search instead for 
Did you mean: 

Problem with sbRIO 9636, status led flashing twice, no connection

Hello,

I have a problem with sbRIO 9636. I tried to change IP address in MAX, and when I tried to save the change, status led started to flashing twice. Now I cannot connect the controller to the PC, it is not visible on the network in MAX. I tried to reset IP, still flashing and not visible...

Can someone help?

0 Kudos
Message 1 of 3
(256 Views)

According to STATUS LED Indicators

Blinks twice and pauses indicate that the sbRIO-9607 is in safe mode. Software is not installed, which is the factory default state, or software has been improperly installed on the sbRIO-9607.

 

Nothing is wrong with the sbRIO.

Try the steps in NI CompactRIO or Network Device Doesn't Show Up or is Missing in MAX

-------------------------------------------------------
Applications Engineer | TME Systems
0 Kudos
Message 2 of 3
(234 Views)

Thank you very much for the prompt response. I tried everything yesterday, but finally I performed IP reset (long press of reset button). After that, it started blinking three times, but no connection. However, after being left on for more than 20 hours, the controller appeared on the network today. Software successfully installed, it works perfectly.

Thanks again!

0 Kudos
Message 3 of 3
(221 Views)