Signal Conditioning
Input Termination For USB-4431
My signal is being driven by a 100 ohm source impedance and is ending at a USB-4431 (USB-4431 Feature Usage - NI).
Should I be terminating at the 4431 with another 100 ohm resistor?
And should I be concerned with the 1k to GND on the pseudo-differential input? That doesn't seem to be able to handle a true differential signal if one side is getting pulled down.
(1,849 Views)
Accepted by topic author kyoung21
the 4431 samples 100kSPS max . For frequencies below 50 kHz you can assume, that the cable is 'just' seen as a capacitance.
The input impedance of the 4431 is 200 kOhm, so check if that load together with your source impedance will hurt you.
Greetings from Germany
Henrik
LV since v3.1
(1,840 Views)
Thank you. Not sure why I was making that more complicated than it needed to be.
No termination will be necessary.
(1,812 Views)