Signal Conditioning
Ni 9237, c series card
Hello
           iam using Ni9237 card and iam using an 100kg load cell My objective is to measure a weight btw 0 - 50 kg.
iexcitation voltage is 5 v. Sensitivity of the load cell is 2mv/v. reading i get for 0 kg is 0.000646431** v/v
and for 50 kg is 0.00116299** v/v . will i be able measure a weight in grams (or atleast 10 grams ) .
** Places of digit i have mentioned in bold are fluctuating  .
(7,776 Views)
Please confirm the Load Cell specifications
Excitation Voltage = 5 V
Rated Output = 2mV/V
Rated Capacity = 100 Kg
100 Kg Load => 10mV output => 10mV/5V = 2mV/V Output
In your setup, a 50 Kg load instead of 0.00116299 v/V = 1.16299 mV/V
So a 1 g load should give you around 20 nV/V output. Lets see whether it is feasible in NI 9237
As per NI 9237 specifications, the Scaling coefficient = 2.9802 nV/V per LSB (ADC = 24 bits = 224 Levels)
So 20nV/V output lies within the detectable range of NI 9237. You should be able to measure the weight in grams.
(7,763 Views)
Hi Ankur,
              The card has a resolution of 24 bits, but the accuracy is 0.038 mv/v. Is it possible then ?. The value which i have got  fluctuations in the later digits is it because of the accuracy.
(7,759 Views)
Accepted by topic author Rameshs
Yes, the variation in measurement may be primarily due to accuracy.
I think that load cell specifications contain accuracy details. Generally it is in % of maximum rated weight.
Trust that helps
(7,756 Views)
Good day
I am to write a code that will read data from two NI 9237 modules (6 channels in total). the modules are to read data from a force balance (strain due to force).
I know its not a matlab community but I could get help, how do I convert strain result to force using a loadcell of specification stated below and the NI 9237. being specific, I need the coefficient to convert strain value given by the daq to force.
Load Cell: STA-7 Aluminum Alloy S Type Low Range Tension/Compression Load Cell
Rated load (kg)                      1, 2, 5, 10, 25, 50
Ultimate breaking load         300% of rated load
Maximum transverse load   <50% of rated load
Output/ Sensitivity                2mV/V (±0.1%)
Zero balance              <±1% of rated load
Excitation voltage                  10vdc recommended, 15vdc maximum
ch6.BridgeMode = 'Full';
ch6.ExcitationVoltage = 10;
ch6.NominalBridgeResistance = 350;
(3,175 Views)