05-12-2020 08:25 PM
Hi! I'm using an NI PXI-5422 Arbitrary Waveform Generator to send a communications signal through an experimental setup (with transducers) to an NI PXI-5124 Digitizer. When I set the input impedance of the digitizer to 50Ω, which is recommended here for "better" communications signals, I don't see the signal at the digitizer. When I set it to 1MΩ, I see the signal but with an additional 60 Hz component. I could set it to 1MΩ and filter out 60 Hz, but I don't want to possibly affect the quality of the signal with the reflections from the 1MΩ circuit path components. So, what could be the reason that I do not see the signal through the 50Ω path and how can I fix it?
05-13-2020 01:59 AM - edited 05-13-2020 02:03 AM
There is a (small) chance that your 50 Ohm resistor in the 5124 has changed his value .. (to short).
EDIT: First test: Use the 5422 (and the NI-FGEN Frontpanel) , directly connect it to the 5124 and use the NI-Scope frontpanel to read the values (and switch input impedances).
Another test: Use a DC voltage source (1.5V battery) and connect it to the input in question.
Use NI-Scope Frontpanel (or MAX test ) to read the DC voltage. (you can add a measure mean voltage 😉 ) .
Switch the input impedances ... no or only minor change in voltage should be observed.
Now put a 1M ohm or a 50 Ohm (2x100Ohm) resistor in series to the voltage source. if series resistor and input impedance match, you should read about 1/2 the voltage.(And you can calculate the exact input impedance, if you know ypour resistors and the voltage source)
If this test runs fine, I would expect a buggy setup in your vi.
05-13-2020 09:47 AM
Thanks Henrik! I tried the first test and I sent 1MHz 1Vpk-pk sine wave from the generator to the digitizer directly.
From CH1 of the digitizer, I got
and from CH0, I got
I don't have the necessary equipment to perform the second test at the moment, but I believe from this test alone, something seems sketchy. Do you think it's 50 Ohm being changed or something else?
05-14-2020 05:26 AM - edited 05-14-2020 05:28 AM
I assume the unit has passed a selftest and autocalibration... .. looks like a defective unit to me.
We just had a 5922 scopecard with (ESD?) blown input repaired...