11-29-2019 02:28 PM
The SCXI-1122 with W1 at (factory setting) and W2 (pullup).
In using a combination of a thermocouple and 0-10 / 0-5 vdc at the SCXI-1322 wired inputs on CH0-CH7.
All are set to differential, 4khz and looking at the physical input it seems stable, at least not to match the data seen in attachment.
It appears as a sensitivity issue in this case, rather than noise. I'm unfamiliar with the SCXI and using multiple type inputs on the same module.
12-02-2019 10:43 AM
the attached signal display (noise) is present on all channels when no change to the voltage input signals are present present.
I mis-spoke about the 4-20ma input. There is a 250 ohm at the device and NIMax is set to read 1-1-5vdc.
when the level of the signal changes, the input appears to change appropriately without the noise amplitude, as if common is missing.
Thank you all for any suggestions!
12-17-2019 01:19 PM
How do i adjust signal sensitivity, in NiMAX?
Verified that noise is not sourced from field wiring by connecting a battery box w/short leads to the input channel.
thanks