06-07-2016 05:05 AM
I have a SCXI-1102 and 100 OHM platinum RTDs.
They are conncted to a drive, if the drive is off, they read the correct temperature but when the drive is enabled the values jump about and are not constant.
I was just wondering if there is something wrong with the setup or could it be some kind of noise.
See attachement for setup.
06-07-2016 09:05 AM
Try a samplerate >300Hz and mean the values of one second ...
Have a look at one RTD signal with the highest possible samplerate, you migth find that there is a lot of power line noise (and higher frequencies if a electronic converter is used)
06-20-2016 10:19 AM
Henrik,
Thanks for the prompt reply.
I am sampling at 1000 samples a second, so I don't think it is that butI have found some more information:
Looking through the documentation of the SCXI-1102 it tells me that I also need a SCXI-1581.
Am I correct or do you know different.
TimS