I'm generating a retriggerable analog output signal, and so I'm using a counter as the sample clock (see: Retriggerable AI Using Retriggerable Counter). I am finding that, above a certain number of samples, and only for certain values of the number of samples, the counter task gives me error -200305, "Desired finite pulse train generation is not possibe." The error crops up only when actually starting the task.
The analog signal that I'm trying to generate will be about 800 kHz, so my counter is set to run at the same frequency. I find that the counter task works fine if the number of samples to generate is anywhere between zero and 671,088 samples. Setting the number of samples to 671,089 gives the error above, as does 671,090 samples and so on. However, using 671,096, the counter task works fine. After that, the counter seems to output fine only if the number of samples is divisible by 8.
The only thing I can think of is that (617088 samples) / (800000 Hz) = 0.839 s. At the internal clock rate of 20 MHz, 0.839 s is 2^24 samples, and it is a 24-bit counter on this hardware. So if it's this internal counter rolling over, that's fine and I can work around that. But if that's the case, what I don't understand is why increasing the number of samples in increments of 8 samples still works.
The hardware is a PXI-6733 board, running with LabView 7.1.1 and NI-DAQmx 8.1.
05-17-2006 05:42 AM
Internal count roll-over sure *sounds* like the most likely culprit. My only guess about the multiples of 8 would be that DAQmx is getting smart behind your back.
When requesting a duration that would take > 2^24 cycles of the 20 MHz timebase, perhaps it automatically chooses to use the 100 kHz timebase instead. Then, it only allows durations that exactly fit into full periods of the 100 kHz timebase. Since you're generating a 800 kHz pulsetrain, you get an exact fit only when the # samples is a multiple of 8. Just a guess, but I've got myself mostly convinced.
If you really need >2^24 pulses, and have the # NOT be a multiple of 8, post back. There *may* be some sneaky tricks if you do all your own manual counter configuration rather than leaving it up to DAQmx...
-Kevin P.
05-17-2006 10:10 AM
05-17-2006 11:52 AM
Hmmm, multiples of 50 & 100? Now I'm puzzled again.
Here's how to make sense of the 100 kHz timebase idea though, even if it turns out not to be the right explanation. For a retriggerable finite pulse train, you actually use a pair of counters. If you were to program it manually, you could set your output counter to generate a continuous pulsetrain at 800 kHz using the internal 20 MHz timebase. This output counter would also be configured to use the other counter's output as a digital level-based pause trigger. So the 800 kHz pulsetrain is only output while the other counter's output is, say, high.
The other counter is configured for retriggerable pulse generation. The pulse duration or high time should be set for (# pulses) / (800e3 pulses/sec). This other counter can be configured to use the 100 kHz timebase, so its high time would then have to be an integer multiple of 10 usec.
So let's see... An 800 kHz pulsetrain is possible with a 20 MHz timebase (exactly 25 cycles). A 700 kHz (28 + 4/7 cycles) or 900 kHz (22 + 2/9 cycles) is not. So when you request those other frequencies, you actually get a near approximation. I dunno if DAQmx can be queried for the actual value correctly or not -- I recall an early version that reported back whatever freq you had asked for rather than what it actually used. Queries based on ticks (rather than time or freq) did return what was actually used, as I recall.
Let's suppose a request for 700 kHz gets truncated to 28 cycles of the 20 MHz timebase making a 1.4 usec period. Then 50 of those periods becomes 70 usec, which is evenly divisible by the 100 kHz timebase. Bingo! (Note: 70 is the least common multiple of 10 and 1.4)
Now suppose the request for 900 kHz turns into 22 cycles of the 20 MHz timebase, or a 1.1 usec period. Now it takes 100 of those periods to get to 110 usec, which is also evenly divisible by the 100 kHz timebase. Bingo again! (Note: 110 is the lcm of 10 and 1.1).
Did you follow the method here? It should help you figure out expected results for various output freqs and #'s of samples.
-Kevin P.